IEPE-2G05 slide

Buy VT IEPE-2G05 (US$399.95)

PC USB 24-bit dual-channel 48 kSPS 0.03Hz~23kHz IEPE Data Acquisition Interface

Introduction

VT IEPE-2G05

VT IEPE-2G05 is a 24-bit, dual-channel, USB data acquisition interface specially designed for use with IEPE sensors such as IEPE accelerometers, IEPE microphones and IEPE hydrophones. Each channel has six calibrated voltage measurement ranges for easy calibration to sensor sensitivity: ± 250 mV, ± 500 mV, ± 1 V, ± 2.5 V, ± 5 V, ± 10 V, and a built-in 24V 4mA current source to drive an IEPE sensor directly. It contains a hardware high pass filter with 5 selectable -3dB corner frequencies: None (0.03Hz), 1.8 Hz, 119 Hz, 236 Hz, 464 Hz. The signals sensed by the IEPE sensors can be amplified and output directly from the stereo headphone jack even without running the PC software. When used in conjunction with the Multi-Instrument® software, the setup allows you to take reliable and quality vibration measurements as simply as plug & play. No external power supply and driver installation is required. It is a truly hassle-free portable noise and vibration measurement solution.

VT IEPE-2G05A

VT IEPE-2G05A has the same specifications as VT IEPE-2G05, except that its Channel B has a 20 dB higher analog gain, with its voltage measurement ranges changed to: ± 25 mV, ± 50 mV, ± 100 mV, ± 250 mV, ± 500 mV, ± 1 V. The increased gain makes it more capable of measuring low level signals such as a low-dBSPL sound.

VT IEPE-2G05B

VT IEPE-2G05B has the same specifications as VT IEPE-2G05, except that its two input channels A & B have a 20 dB higher analog gain, with their voltage measurement ranges changed to: ± 25 mV, ± 50 mV, ± 100 mV, ± 250 mV, ± 500 mV, ± 1 V. The increased gain makes it more capable of measuring low level signals such as a low-dBSPL sound.

VT IEPE-2G05C/D/E

VT IEPE-2G05C/D/E have the same specifications as VT IEPE-2G05, except that their Channels B have a 40 dB higher analog gain, with their voltage measurement ranges changed to: ± 2.5 mV, ± 5 mV, ± 10 mV, ± 25 mV, ± 50 mV, ± 100 mV. The increased gain makes them more capable of measuring extra low-dBSPL sound. Their Channels A have a 0dB /20dB / 40dB higher analog gain respectively than that of VT IEPE-2G05.

All the above models can be switched to a voltage measurement mode (e.g. for the purpose of accepting the voltage pulses output by a non-IEPE tachometer) by cutting off the 24V 4mA constant current supply through an internal DIP switch in each channel. The 4mA constant driving current can be modified as well (e.g. for the purpose of driving a very long cable). These customizations are recommended to be done in the factory upon request before shipping.

It is possible to run multiple VT IEPE-2G05/A/B/C/D/E using multiple instances of the software on the same computer.
Find out more: Manual

Package Contents

  1. VT IEPE-2G05/A/B/C/D/E unit with a hardware activated Multi-Instrument Standard software license
  2. USB cable (1.5 m)
  3. CD (contains the copy-protected Multi-Instrument Software)
  4. Carrying case

Add-on Items

  1. IEPE Accelerometers, IEPE Velocity Sensors, IEPE Impedance Heads, IEPE Impact Hammers (including a mounting stud / screw and a cable from sensor to BNC, default: 2m)
  2. IEPE Measurement Microphones (including a small mic stand, a mic clip and a cable from mic to BNC, default: 2m)
  3. IEPE simulators
  4. Magnetic Mounting Bases
  5. Software License Upgrades
  6. USB Isolators

Note: It is possible to place most add-on items inside the same carrying case with VT IEPE-2G05/A/B/C/D/E.

Hardware Specifications

General
Number of Input Channels 2
Sampling Frequency 48 kHz (original), 44.1 kHz, 32 kHz, 22.05 kHz, 16 kHz, 11.025 kHz, 8 kHz, 4 kHz, 2 kHz ...
ADC Bit Resolution 24 Bits (can be reduced to 16 bits or 8 bits)
Input Voltage Ranges IEPE-2G05:
± 250 mV, ± 500 mV, ± 1 V, ± 2.5 V, ± 5 V, ± 10 V
IEPE-2G05A:
Ch.A: ± 250 mV, ± 500 mV, ± 1 V, ± 2.5 V, ± 5 V, ± 10 V
Ch.B: ± 25 mV, ± 50 mV, ± 100 mV, ± 250 mV, ± 500 mV, ± 1 V
IEPE-2G05B:
± 25 mV, ± 50 mV, ± 100 mV, ± 250 mV, ± 500 mV, ± 1 V
IEPE-2G05C:
Ch.A: ± 250 mV, ± 500 mV, ± 1 V, ± 2.5 V, ± 5 V, ± 10 V
Ch.B: ±2.5 mV, ±5 mV, ±10 mV, ±25 mV, ±50 mV, ± 100 mV
IEPE-2G05D:
Ch.A: ± 25 mV, ± 50 mV, ± 100 mV, ± 250 mV, ± 500 mV, ± 1V
Ch.B: ± 2.5 mV, ± 5 mV, ± 10 mV, ± 25 mV, ± 50 mV, ± 100 mV
IEPE-2G05E:
± 2.5 mV, ± 5 mV, ± 10 mV, ± 25 mV, ± 50 mV, ± 100 mV
Input Connectors & Interface BNC, Single Ended, IEPE, 24V 4mA
Input Coupling Type IEPE-2G05:
AC (High pass filtered at 0.03Hz)
IEPE-2G05A:
Ch.A: AC (High pass filtered at 0.03Hz)
Ch.B: AC (High pass filtered at 0.3Hz)
IEPE-2G05B:
AC (High pass filtered at 0.3Hz)
IEPE-2G05C:
Ch.A: AC (High pass filtered at 0.03 Hz)
Ch.B: AC (High pass filtered at 3 Hz)
IEPE-2G05D:
Ch.A: AC (High pass filtered at 0.3 Hz)
Ch.B: AC (High pass filtered at 3 Hz)
IEPE-2G05E:
AC (High pass filtered at 3 Hz)
Input Isolation No (Isolation can be achieved through a USB isolator)
Input Impedance IEPE-2G05:
510 kΩ
IEPE-2G05A:
Ch.A: 510 kΩ
Ch.B: 51 kΩ
IEPE-2G05B:
51 kΩ
IEPE-2G05C:
Ch.A: 510 kΩ
Ch.B: 5.1 kΩ
IEPE-2G05D:
Ch.A: 51 kΩ
Ch.B: 5.1 kΩ
IEPE-2G05E:
5.1 kΩ
Input High Pass Filter None, 1.8 Hz, 119 Hz, 236 Hz, 464 Hz
Frequency Response IEPE-2G05:
0.03 Hz ~ 22.8 kHz
IEPE-2G05A:
Ch.A: 0.03 Hz ~ 22.8 kHz
Ch. B: 0.3 Hz ~ 22.8 kHz
IEPE-2G05B:
0.3 Hz ~ 22.8 kHz
IEPE-2G05C:
Ch.A: 0.03 Hz~22.8kHz
Ch.B: 3 Hz~22.8kHz
IEPE-2G05D:
Ch.A: 0.3 Hz~22.8kHz
Ch.B: 3 Hz~22.8kHz
IEPE-2G05E:
3 Hz~22.8kHz
Frequency Accuracy 50 PPM
Anti-aliasing Filter 22.8 kHz at Sampling Rate 48 kHz, proportionally adaptive to Sampling Rate Chosen
Buffer Size Virtually unlimited (streaming mode)
Voltage Accuracy ±0.5% at 1 kHz
Output Connector & Interface ϕ3.5 mm Stereo Audio Jack
Output Voltage Range ±0.5 V
Digital Input / Output Standard USB Audio Class 1
Calibration Individually done at factory, user re-calibratable
PC Interface USB 2.0 Full Speed / USB 1.1
Device Category in Multi-Instrument ADC Device Sound Card MME
DAC Device Not Applicable
Power Bus powered by USB port, no external power source required
Power Consumption Max. 0.5W
Dimensions 128 mm (L) x 57 mm (W) x 24 mm (H), anodized aluminum case
Operating Temperature 0° C ~50° C
System Requirements Windows XP, Vista, 7, 8, 8.1, 10, 11, 32 bit or 64 bit
Frequency Response of Built-in High Pass Filter (sampled at 48 kHz)

Frequency Response of Built-in High Pass Filter (sampled at 48 kHz)

Examples

VT IEPE-2G05 with Multi-Instrument Pro + Spectrum 3D Plot
VT IEPE-2G05 with Multi-Instrument Pro + Spectrum 3D Plot
VT IEPE-2G05 with Multi-Instrument Pro + Spectrum 3D Plot + Vibrometer
VT IEPE-2G05 with Multi-Instrument Pro + Spectrum 3D Plot + Vibrometer
Half-sine Shock Test
Half-sine Shock Test (Video: ShockTestWithMulti-Instrument.mp4)
Sound Quality Measurement
Sound Quality Measurement: dBSPL, Loudness, Loudness Level, Sharpness, Articulation Index and Open Articulation Index of a white noise in a Free Field with an IEPE Measurement Microphone (Note: Multi-Instrument Pro or above is required)
Absolute Acceleration Shock Response Spectrum of a Half-sine Pulse
Absolute Acceleration Shock Response Spectrum of a Half-sine Pulse (Multi-Instrument Pro or above is required)
Relative Velocity Shock Response Spectrum of a Half-sine Pulse
Relative Velocity Shock Response Spectrum of a Half-sine Pulse (Multi-Instrument Full Package is required)
Relative Displacement Shock Response Spectrum of a Half-sine Pulse
Relative Displacement Shock Response Spectrum of a Half-sine Pulse (Multi-Instrument Full Package is required)
Pseudo Velocity Shock Response Spectrum of a Half-sine Pulse
Pseudo Velocity Shock Response Spectrum of a Half-sine Pulse (Multi-Instrument Full Package is required)
Equivalent Static Acceleration Shock Response Spectrum of a Half-sine Pulse
Equivalent Static Acceleration Shock Response Spectrum of a Half-sine Pulse (Multi-Instrument Full Package is required)
Damping Ratio and Q factor Measurement using Free Vibration Decay method and Half-Power method
Damping Ratio and Q factor Measurement using Free Vibration Decay method and Half-Power method (Note: Multi-Instrument Full Package is required)
Sound Intensity Measurement
Simultaneous Sound Intensity and Sound Pressure Measurement using a P-P Sound Intensity Probe (Note: Multi-Instrument Full Package is required)