Can't measure IMD of both channels using DS02810

PostPosted: Fri Mar 03, 2017 12:06 pm
by bhands5
Hi,

I am trying to measure IMD of both channels using DSO2810 and Multi Instrument 3.6 version on Windows 7.
But it shows values only for Channel A in Spectrum Analyzer, for Channel B only Peak Frequency (f1) is shown.

is this a known limitation?

Regards,
Santosh

Re: Can't measure IMD of both channels using DS02810

PostPosted: Fri Mar 03, 2017 7:39 pm
by VirtinsTech
Thank you for your questions.

For software versions before 3.6.0.4, IMD measurement requires the signal generator of the software to generate a multi-tone signal consisting of two test frequencies. If the generated signal is mono, then only the IMD of Channel A is measured. If the generated signal is stereo, then the IMDs of both channels are measured. For DSO-2810, the generated signal is mono.

In the latest version (V3.6.0.4), the above restriction has been removed. Now, IMD can be measured even if the signal generator is not running, therefore an audio CD containing the IMD test tones can also be used in IMD tests instead of the software's signal generator. The software will detect the two IMD test frequencies automatically. You can download the latest version at: www.virtins.com/MIsetup.exe .