strange results from THD measurement of mains

PostPosted: Tue Nov 28, 2017 1:58 am
by pete_dl
I have followed the video on mains THD measurement. I used the ground differential measurement method on the wall outlet in the UK.

It reports the largest frequency being at 25hz and sometimes jumping to 34hz. This makes no sense to me, as I would have thought that a distorted 50hz sine wave would only have harmonics above 50hz.

I have attached a screen grab.

Have I made a configuration mistake ?

Regards,
Peter.
MIgrab3.jpg
MIgrab3.jpg (250.75 KiB) Viewed 46729 times

Re: strange results from THD measurement of mains

PostPosted: Thu Nov 30, 2017 11:50 pm
by VirtinsTech
Does this problem exit in Earth Grounded Single End Measurement?

Re: strange results from THD measurement of mains

PostPosted: Sun Dec 03, 2017 7:07 am
by pete_dl
I have now tried Earth Grounded Single Ended Measurement, and it still says the peak frequency is 25hz

Re: strange results from THD measurement of mains

PostPosted: Sun Dec 03, 2017 8:41 am
by pete_dl
I have also tried with my own generated sine wave. I created a WAV file with a 100hz pure sine wave and verified it with a frequency analysis app. I then played it through a DAC and into the DSO-2810R. The MI software reports a peak at 50hz and sometimes jumps to 68hz.

Re: strange results from THD measurement of mains

PostPosted: Mon Dec 04, 2017 8:43 pm
by VirtinsTech
Did you see any abnormal in the captured waveform in the oscilloscope?

For example, from the screenshot you provided in the first post, the waveform in the oscilloscope did show a frequency of about 34 Hz. To check whether the DSO hardware is faulty or not, you can use the DSO to measure the signal output from an external signal generator.

Re: strange results from THD measurement of mains

PostPosted: Mon Dec 04, 2017 11:28 pm
by pete_dl
the waveform looks fine. I have independently confirmed the input signal is 100hz sine wave.

What the device seems to report is half the correct frequency and sometimes 0.68 of the correct frequency.

When I drive it with 100hz, it reports either 50hz or 68hz

when I drive it with 50hz UK mains, it reports either 25hz or 34hz. 34hz is 50hz * 0.86

What it feels like is the ADC sampling clock is flicking between two incorrect values.

Re: strange results from THD measurement of mains

PostPosted: Wed Dec 06, 2017 12:01 am
by VirtinsTech
If that is the case, please contact support@virtins.com if you need to send it back for repairing.